here is an example on 2 different micrographs, my training set contains ~10000 particles and I have used 500 particles per micrograph, 150A diameter, the autopick FOM here is set to 0 |
Beta Was this translation helpful? Give feedback.
-
Have you tried adjusting the likelihood threshold for selecting particles (I think FOM in the Relion interface)? Setting this to a larger number will reduce false positives and could be a way to remove the carbon picks. When you run extract, I would recommend setting the particle size to that of your largest particle. However, topaz is generally good at localizing the centers, so it tends to work with smaller numbers as well. If you see that topaz is producing multiple picks from a single large particle, then the size should be increased. On the other hand, if you have closely packed particles, increasing the extraction radius size means that you may end up missing some of them, because they are eliminated when their neighbor is selected. The expected number of particles should be the average number of particles per micrograph, so, yes, 400. Other than the picks over carbon, are there any clear issues with these picks? It seems like raising the likelihood threshold is probably the easiest and most fruitful thing to try. |
Beta Was this translation helpful? Give feedback.
-
Hi,
I am quite new in cryoEM and definitely new in Topaz.
I am trying to use it to pick a multidomain protein 150KDa in Relion4.
I have some questions I would like to ask to Topaz specialists.
Thanks a lot for your help,
GIA