QCoDeS 0.63.0 (2026-11-11)
Breaking Changes:
Several instrument driver parameters now return numbers or booleans instead of the raw string
from the instrument, or an integer for an on/off state. Code that compares the returned value to
a string, or to
0/1by identity, needs updating:Keithley2450.line_frequencyandKeithley2450Sense.countnow returnint, andKeithley2450Sense.nplcandKeithley2450Source.delayreturnfloat.Keithley7510Sense.user_delaynow returnsfloat, andKeithley7510Sense.average_count,Keithley7510DigitizeSense.acq_rateandKeithley7510DigitizeSense.countreturnint.SR86xBuffer.count_capture_kilobytesnow returnsint.RigolDG1062Burst.ncyclesnow returnsint, andRigolDG1062Burst.phase,RigolDG1062Burst.time_delayandRigolDG1062Channel.duty_cyclereturnfloat.RigolDS1074Z.trigger_levelandTektronixTPS2012.trigger_levelnow returnfloat.TektronixDPOHorizontal.record_lengthnow returnsintinstead offloat, as therecord length is a number of samples.
AgilentE8267C.pulse_widthnow returnsfloatand its unit issinstead ofns,matching the value sent to and returned by the instrument.
RohdeSchwarzRTO1000ScopeChannel.impedanceandKeysightE4980A.rangenow returnfloat.RohdeSchwarzZNBChannel.bandwidthnow returnsfloat, e.g.10000.0instead of10000.stateanddirect_outputand the instrumentDC_outputofTektronixAWG5014now returnbool. They acceptTrue/False,1/0and"on"/"off".TektronixAWG5014.sequence_posnow returnsint.KeysightB1520A.ac_dc_volt_monitornow returnsbooland also accepts1/0and
"on"/"off". Add types to generic instrument parameteres #8658The field units of an
AMIModel430can no longer be changed while it is part of anAMIModel4303D, and doing so now raises aRuntimeError. The set point and field limit ofthe 3D driver are in the field units of the magnets, so changing them silently changed their
meaning by a factor of 10. Set the field units before creating the 3D driver, or close the 3D
driver first.
The coordinate vectors of
AMIModel4303D(cartesian,spherical,cylindricalandtheir
*_measuredversions) now return a named tuple (CartesianVector,SphericalVectororCylindricalVectorfromqcodes.math_utils) instead of a list, sothe components can also be accessed by name, e.g.
cartesian_measured().z. Code that comparesthe result to a list, or modifies it, needs updating. Setting these parameters still accepts
lists, tuples and numpy arrays. Improve Ami430 safety and type annotations #8679
The
auto_zero_onceparameter ofKeithley7510Senseand thetrigger_in_ext_clearparameter of
Keithley7510have been replaced by methods of the same name. Both werecommand-only parameters, so calling them without an argument, e.g.
dmm.sense.auto_zero_once(), only returned the cached value and did not send anything to theinstrument. Calling them like this now sends the command. Code that passed a value to them, e.g.
dmm.trigger_in_ext_clear(None), must drop the argument, and they are no longer part of theinstrument's parameters or snapshot. Add type variables to parameters in several drivers #8694
The
trigger_countparameter ofKeithley6500, which sets the number of times a scan isrepeated (
ROUT:SCAN:COUN:SCAN), now accepts the full range documented in the DMM6500Reference Manual: 0 to 100,000,000, where 0 repeats the scan until it is aborted. It previously
only accepted 1 to 9999 or the strings
"inf","default","minimum"and"maximum",which were copied from the Keithley 2000 and are not documented for this command. These strings
are now rejected; use 0 instead of
"inf". Whole-number floats are accepted and sent asintegers, and the parameter is annotated as
Parameter[int, Self]. Add type variables to parameters in several drivers #8694Improved:
RuntimeError: dictionary changed size during iterationif the garbage collector cleaned upanother instrument at the same time. The instrument then stayed registered, so its name could
not be reused. Fix race between closing instruments and garbage collection #8680
the type of its value, e.g.
Parameter[float, Self], and all of its example drivers areannotated this way. This also fixed bugs in the example code: the Ithaco example passed
unitsinstead of
unit, the Keithley 2600 example lacked itsmodelattribute and_display_settextmethod, and the Alazar example used outdatedTraceParameterarguments. Add type variables to parameters in several drivers #8694transaction, inserting the snapshot after the other run metadata updates to
avoid repeatedly rewriting large snapshots. The complete snapshot is committed
before background writers and pending subscribers are started.
Improved Drivers:
Add convenience methods to Tektronix DPO7200xx (
single,download_waveforms,get_timebase) and AWG70000A (upload_seqx) drivers for streamlined waveform acquisition and sequence upload workflows. Tektronix DPO7200xx and AWG70000A — Convenience Methods for Waveform Workflows #7865The current ranges of the Keithley 2634B, 2635B and 2636B include 10 mA; their lists contained 10 µA twice in its place, so a 10 mA range was rejected. Fix the 10 mA current range of the Keithley 2634B, 2635B and 2636B #8618
Most instrument driver parameters are now annotated with the type of their value, e.g.
Parameter[float, Self], so static type checkers can infer the type returned bygetand accepted by
set. As part of this, the mock instruments no longer pass the defaultparameter_class=Parameterexplicitly, andKeysight34980A.moduleis typed as a mappingfrom slot number to switch matrix submodule.
Fixed the following driver bugs found along the way:
RigolDS4000.enable_auto_scalealways read back asTrue, sincebool("0")isTrue, and was set withTrue/Falseinstead of1/0.RohdeSchwarzRTO1000.trigger_modequeried the trigger source instead of the trigger mode.Keithley6500.digitssent a set command with a stray?.MessageBuilder.pch_queryof the Keysight B1500 sentPCHinstead ofPCH?.LakeshoreModel325Heater.output_rangeaccepted the high (25 W) range on loop 2, which isonly supported on loop 1.
LakeshoreModel325Sensor.curve_indexnow accepts0(no curve), which the instrumentalso reports when the selected curve does not match the input, and rejects non-integer values.
Whole-number floats such as
23.0are still accepted and sent as integers.LakeshoreModel325curve list now includes user curve 35, soupload_curve(35, ...)no longer raises an
IndexError.LakeshoreModel325Curve.get_datanow reads all 200 points of a curve instead of 199.LakeshoreModel325Curve.set_datanow writes curve points with 6 significant digits, theresolution of the instrument, instead of rounding them to 3 decimals.
LakeshoreModel325Sensor.curvenow raises aRuntimeErrorwhen no curve is selected forthe sensor, instead of returning a curve 0, which the instrument does not have.
RohdeSchwarzZNBChannel.bandwidthfailed to read a non-integer bandwidth such as 1.5 Hz,which the instrument supports.
current_complianceandpower_complianceof theKeysightB1517Asweep could notbe set back to
None, which leaves the compliance out of theWVcommand. Since a powercompliance can only be sent together with a current compliance, clearing the current
compliance also clears the power compliance.
Keysight34980A.scan_slotscan now be called again; it skips slots that already have amodule instead of failing to register a duplicate submodule.
Keithley7510DigitizeSense.aperturewas labelled inusalthough the instrument usesseconds; its unit is now
s.averages,pointsandactive_traceof the Keysight PNA (KeysightPNABase) driversnow reject non-integer values. Previously
averagesfailed to format a float such as2.5and the others sent it to the instrument. Whole-number floats such as2.0arestill accepted and sent as integers. Add types to generic instrument parameteres #8658
CopperMountainM5xxx.averages,AimTTiChannel.curr_range,RigolDG1062Burst.ncycles,Keithley7510Sense.average_countand thedwell,pause,curve_numberandexcitation_range_numberofLakeshoreModel372Channelnow reject non-integer values.Previously these were sent to the instrument, or for
CopperMountainM5xxx.averagessilentlytruncated. Whole-number floats such as
2.0are still accepted and sent as integers.Also removed the incorrect unit of
AimTTiChannel.curr_range, which is a range number, fixedthe label of
LakeshoreModel372Channel.temperature_coefficient, and removed incorrect unitsfrom
RigolDS1074Z.trigger_mode,waveform_npoints,waveform_yoriginandwaveform_yref. Switch float to permissive int for parameter validators and add types #8678The field parameters of
AMIModel4303D, such ascartesian,xandfield_measured,now use the field units of the individual magnets, tesla or kilogauss, instead of always
reporting tesla. A warning is logged if the magnets use different field units. Also removed the
placeholder unit
tuple?ofAMIModel4303D.sphericalandcylindrical, whose componentshave different units, which are now described in their docstrings. Closing an
AMIModel4303Dnow detaches its magnets, soset_field,zeroand changing the units ofa magnet that is still open no longer fail. Improve Ami430 safety and type annotations #8679
The parameters of the Basel SP983, SP983a and SP983c, Ithaco 1211, Keithley 2450, DMM7510, S46
and 2600 drivers that were not yet typed are now annotated with the type of their value, e.g.
Parameter[float, Self], so static type checkers can infer the type returned bygetandaccepted by
set. Add type variables to parameters in several drivers #8694New:
CartesianVector(x,y,z),SphericalVector(
field,theta,phi) andCylindricalVector(rho,phi,z) toqcodes.math_utils, for the components of a field vector in a given coordinate system. 8679with qcodes.parameters.record_parameter_accesses() as accesses:block,every
getandsetof a parameter made by the current thread (or asyncio task) is appended toaccessesas aqcodes.parameters.ParameterAccess: the names of the parameter and its instrument, the value and raw value, when theaccess started, how long it took and the error it raised, if any. A set that ramps in steps is recorded once per step.
Accesses are appended to a new list, or to a mutable sequence that is passed in, whose
appendmust only store them.qcodes.parameters.log_parameter_accesseslogs the accesses to theqcodes.parameters.parameter_accessloggerinstead. Recording only stores data: it is not a way to react to an access, and a record holds names rather than the
parameter or instrument. This makes it possible to see all instrument communication made by a piece of code without
changing it. 8684
Under the hood:
instruments no longer first try to connect to a real instrument, and a number of
unneeded sleeps and waits have been removed or shortened. 8725
InstrumentGroupexample notebook ramps its gate in fewersteps, and the
qcodes.sphinx_extensions.parse_parameter_attrSphinx extension now parses thesource of each class only once instead of once for every attribute documented on it. 8742
root, so setting
PYTEST_DEBUG_TEMPROOT(or--basetemp) relocates all of them. 8749PRAGMA synchronous=OFFso that commits are not flushed to disk.This discussion was created from the release QCoDeS 0.63.0.