Hi @soyyso! It sounds like the first nose poke in your workflow is incurring some startup cost, which can happen when operators first initialize. I tried to reproduce it on my end with a simple closed-loop workflow like you described, but I am not seeing that first nose-poke cost, so it might be something in your workflow. Honestly though, I think 10ms might be acceptable given the reaction time for a mouse between mouse poke, valve open, and water delivery. If you want to get rid of it, I guess you can either manually trigger the first nose poke for the experiment and discard it. I can also take a look at your workflow to see the issue. I have some of your earlier versions; I'm not sure if they have been updated? |
If this is your constraint, you will want to move your time-sensitive logic to an embedded system. A general-purpose computer will simply not be able to ever give you this guarantee. |

Uh oh!
There was an error while loading. Please reload this page.
Hello everyone!
I am using a system in which the signal flows from "nose poke → Harp board → PC → valve opening".
For the first nose poke, I observe approximately 11 ms of jitter from nose-poke detection to valve opening. However, for subsequent nose pokes, the jitter becomes stable at around 1–2 ms.
Would an initial jitter of approximately 11 ms be considered negligible for this type of behavioral experiment? Ideally, I would like to maintain the jitter consistently within the 1–2 ms range, including the first nose poke.
Has anyone experienced a similar issue, and is there a way to reduce this initial jitter?