Replies: 2 comments 4 replies
-
|
HU values are not modified beyond the typical loading process. It uses the mean of the two surrounding background ROIs when comparing to the ROI. Any chance the HU values are negative like #539 ? (I've not looked at that one closely yet) |
Beta Was this translation helpful? Give feedback.
-
|
I'm looking at the algorithms Doselab uses and it seems they're all rescaled up by 1000 to every HU, and then reported as percentages. The consistency of noise and agreement with Doselab seems to be improved by rescaling. The area we used to analyze for this purpose was non-traditional and the signal could swing from -50 to +10. What are your thoughts? Has there been any thought in reducing the size of the ROI's for contrast and for HU by some small amount to remove the edges and use median in place of mean? |
Beta Was this translation helpful? Give feedback.
Uh oh!
There was an error while loading. Please reload this page.
-
I'm looking into the CTP515 low contrast visibility section. We're trying to get signal to noise directly to compare with previous tools. Does the algorithm take the HU values without modification or is there a normalization of some kind to account for the scale starting at -1000? Is there a reference for how this decision was made?
Beta Was this translation helpful? Give feedback.
All reactions